I am interested in using a C-type thermocouple for high temperature measurements. Our thermocouple DAQ, Measurement Computing's USB-TC, cannot directly interface with C-type thermocouples (it is not of one of the included settings), but I wanted to try and back-door it with the following method:
Connect the C-type thermocouple to one of USB-TC's thermocouple channels (for example, channel 0)
Set Channel 0's thermocouple type to a thermocouple type with a similar EMF/voltage range as C-types, such as K or J-types (see below figure of thermocouple output voltage to temperature)
Set Channel 0 to read out not as a temperature, but as a voltage reading (I can accomplish this with Measurement Computing's python package, mcculw)
Currently, my main concern is how to account for the reference junction temperature in the conversion of Channel 0's stated voltage and the final C-type thermocouple temperature, assuming the reference junction is at 20 C. From the USB-TC's manual, I understand that the DAQ automatically corrects for the cold junction compensation by subtracting it from the analog input's thermocouple voltage measurement. However, is this correction performed when I set Channel 0 to read out voltages rather than temperatures? If not, how do I calculate the cold junction compensation voltage myself?
Finally, are there other inaccuracies/obstacles that I should consider that might make this C-type measurement impossible/not useful? Thank you in advance for any and all advice.
Question
Lwin D
Hello,
I am interested in using a C-type thermocouple for high temperature measurements. Our thermocouple DAQ, Measurement Computing's USB-TC, cannot directly interface with C-type thermocouples (it is not of one of the included settings), but I wanted to try and back-door it with the following method:
Currently, my main concern is how to account for the reference junction temperature in the conversion of Channel 0's stated voltage and the final C-type thermocouple temperature, assuming the reference junction is at 20 C. From the USB-TC's manual, I understand that the DAQ automatically corrects for the cold junction compensation by subtracting it from the analog input's thermocouple voltage measurement. However, is this correction performed when I set Channel 0 to read out voltages rather than temperatures? If not, how do I calculate the cold junction compensation voltage myself?
Finally, are there other inaccuracies/obstacles that I should consider that might make this C-type measurement impossible/not useful? Thank you in advance for any and all advice.
Link to comment
Share on other sites
5 answers to this question
Recommended Posts
Create an account or sign in to comment
You need to be a member in order to leave a comment
Create an account
Sign up for a new account in our community. It's easy!
Register a new accountSign in
Already have an account? Sign in here.
Sign In Now