ADP3250 Scope Reading With No Probes Connected


janpi

Question

I have recently purchased a new ADP3250. I have noticed that if no probes/inputs are connected the scope still shows 3 mV peak to peak reading of what looks as a random noise. If I connect supplied probe and shorten its ends the reading jumps to approximately 20 mV. Screen shot with shorten probe leads is attached.

Is this normal? Am I doing anything wrong? It seems high to me as I cannot measure any signal below 20 mV noise level.

Thank you for any help.

ADP3250.png

Link to comment
Share on other sites

7 answers to this question

Recommended Posts

  • 0

Thank you for your quick feedback and UI tip Attila.

I had the similar thought so I took the ADP3250 and the battery powered laptop (no charger connected only the ADP3250) to a room with no plugged electronic/electrical devices and unfortunately saw the same output. The laptop MacBook Air M2.

Any thoughts what should I try to troubleshoot?

Link to comment
Share on other sites

  • 0

Hi @janpi

I don't know what could be the cause of this.
You could use Filter channel to filter hf components and/or various averaging options in the application.

image.png

 

Edit: To measure the input 'internal' noise a (50R) terminator or a very short wire shorting the BNC input should be used.
Additional wire will absorb more radiation. For instance a probe ground loop can be used to locate the source/orientation of the radiation:
image.png

Link to comment
Share on other sites

  • 0

Thank you for another tip Attila! There is so much to learn in WaveForms, btw it is a great app.

I made some progress today. I have used different design probe and with shorted end I am getting noise just < 3 mV pk2pk. See screenshot.

Can you please let me know what are you getting on C3 on your ADP3450 with no probes connected?

Screen Shot 2023-10-31 at 4.37.42 PM.png

Link to comment
Share on other sites

  • 0

Hi @janpi

On 10/31/2023 at 11:38 AM, attila said:

Edit: To measure the input 'internal' noise a (50R) terminator or a very short wire shorting the BNC input should be used.
Additional wire will absorb more radiation.

A floating high impedance input is not a good metric, it can pick up random radiation from anywhere.