Daqami and Labview data not matching


Mackenzie Cox

Question

I have several sensors reading in daqami and labview and have had no issues. When I connected my flow sensors, however, the values on daqami look perfectly find, but the values in labview are way way off. This is the same in instacal's "test" feature. I have all of my calibration settings the same in daqami and instacal. Do you know how I can fix this?

Edited by Mackenzie Cox
Link to comment
Share on other sites

2 answers to this question

Recommended Posts

  • 0

With DAQami, you can change the input channel mode in the Channel section. However, LabVIEW is different. It uses InstaCal to configure the channels. Close LabVIEW and run InstaCal. It will prompt you to add the USB-2416 to the board list if it still needs to. Once in the list, double-click it to bring up the Configuration panel. Select each channel on the Analog In Channel tab and set it as it is set in DAQami. Close InstaCal and start-up LabVIEW; it will read InstaCal's configuration file, and your values should not match. 

Link to comment