  1. Hello @Jonaskb. The DT9800 Series driver is 64-bit compatible. The driver works with LabVIEW 2023 Q3 (64-bit) in a Windows 11 64-bit OS. We support using Data Translation (DT####) devices in LabVIEW using Data Translation's LV-Link3 library. We do not have a library for NI DIAdem and have not supported NI DIAdem. Please send me the folder with files and instructions to review. Regards, Fausto
  2. Hello @Joero. Thank you for testing the other OS, systems, and cables. Testing with a Windows 10 system rules out any OS upgrade issue. Testing with Linux rules out an issue with the Windows driver. At this point, it appears that the USB front-end circuitry on the MCC USB-CTR08 may be corrupted. I don't have any other troubleshooting steps and all your results point to non-functional device. I cannot hypothesize what happened to the device since it initially worked back in October 2023. Did anyone else use the device since then? The device does not go bad just by sitting unused for months. Regards, Fausto
  3. @Display Name, What are the outputs when you run the following command with the MCC 152 only, with the other board(s) only, and with all HAT boards mounted on the RPi? i2cdetect -y 1 Thank you.
  4. Hello @Display Name. The MCC 152 uses the I2C interface pins (GPIO 2 and GPIO 3); MCC 152 Electrical Specifications. Other mounted HAT boards cannot use the I2C interface pins if a MCC 152 is mounted on the Raspberry Pi. Regards, Fausto
  5. Hello @jiale. Which software and version was used to create that .dat file? Our DASYLab software application has a module to write data to a DIAdem DAT file, but no option to read the .dat file format. Consider using the NI DIAdem application. Regards, Fausto
  6. Hello @Jonaskb. We do not provide support for NI DIAdem software. On your Windows 11 system, do you have a C:\Program Files (x86)\Data Translation\DeviceDrivers\DT9800 directory with files? With the DT9803 device connected to your Windows 11 system, install Data Translation's QuickDAQ application and then launch it. Does the application detect the DT9803? If so, then add the device to the right-side column and continue with QuickDAQ. Once in the QuickDAQ environment, press the start acquisition or record button to capture data. If that all works, then the issue is not with the DT9803. Regards, Fausto
  7. Hello @WFTXglass. Please provide more information regarding your issue, i.e. screen captures, worksheet, any processes running in the OS background, any virus scan apps running, are hibernation and sleep modes disabled. Have you attempted to create a simple worksheet to debug your issue? Regards, Fausto
  8. Hello @WFTXglass. You may be using an older version of InstaCal, not supported under Windows 11. Please download and install the current version of InstaCal. Restart your Windows 11 system and then retest the MCC USB-1608X-2AO with InstaCal first. If the device is detected by InstaCal and no errors, then close the application and launch DASYLab 2016. Any errors using the device with your DASYLab worksheet? If so, create a new simple worksheet to test the MCC USB-1608X-2AO. Reply with your script and screen captures if errors continue to be an issue. Regards, Fausto
  9. Hello @shebbar. Which OS and software application are you using with your MCC USB-1608G devices? Regards, Fausto
  10. Hello @Joero. Thank you for the update and screen captures. Your serial number 1FC83C9 is out of warranty (shipped Sept. 2021) and the MCC USB-CTR08 is not cost-effective to repair. You mentioned that you already tested the device with other Windows systems and USB cables. Try testing a different PC and Windows OS if all previously tested systems were the same model and running the same OS version. Also, try using other cables, ports, and USB hubs. Regards, Fausto
  11. Hello @Joero. Which version of Windows are you using? Download, install, and launch MCC's InstaCal utility to detect the MCC USB-CTR08 connected to the target system. File location: https://files.digilent.com/#downloads/InstaCal/ Is the device detected by InstaCal? What other software do you use with this device? What is the device's serial number on the backside label (or in InstaCal)? Regards, Fausto
  12. Hello @gilgeonics. Please download and install Data Translation's OMNI software. Data Translation devices (i.e. DT9816) are not compatible with MCC's InstaCal software. Regards, Fausto
  13. Hello @Ron Byron. A PDF version of the MCC Universal Library (UL) Help is not available. Please refer to the online UL Help for Windows. Regards, Fausto
  14. Hello @nicktts. If you recently installed DAQami on a Windows 11 system (or upgraded to Windows 11), then you need to update the target system with the latest version of InstaCal. https://files.digilent.com/downloads/InstaCal/icalsetup.exe DAQami installs a version of InstaCal which is not compatible with Windows 11. Please confirm that the target system has the current version (6.74) of InstaCal installed. If the system does have the latest version of InstaCal installed, then it could be a corrupt daqami.mcfg configuration file. Browse to the C:\Users\{username}\Documents\Measurement Computing folder. Delete the daqami.mcfg and restart DAQami. DAQami will create a new configuration file. Regards, Fausto
  15. Hello @daquser. The DT9837 Series (i.e. DT9837A and DT9837B) supports a ±10 V signal on the analog input channels. Have you tested the daq and sensor combination already, but noticed an issue? Please provide a datasheet of your sensor/probe. Regards, Fausto
  16. @Mike R, Thank you for correcting the NI board model as a NI PCI-6035e. If you are only using the analog input channels, then the MCC USB-1608G does have nearly compatible specifications. The MCC USB-1608G does not have an input voltage range setting of +/-500 mV nor +/-50 mV. MCC USB-1608G NI PCI-6035e Regards, Fausto
  17. Hello @lesthaeghet. Thank you for testing. Please test a few other channels with TB3, 18 (5V Analog) and in single-ended mode. If you continue to get similar readings, then please reply via private message with the serial number from the backside label on the DT9844; an image of the label is helpful. The serial number is used to determine if the module can be serviced. Regards, Fausto
  18. Hello @Osiris. Please confirm the SPI and I2C options are selected under Preferences - Raspberry Pi Configuration. Also, since you changed your board stack, you must update the saved EEPROM images so that the library has the correct board information. You can use the DAQ HAT Manager or the Terminal command: sudo daqhats_read_eeproms If the MCC 152 continues to be undetected, remove both boards and install just the MCC 152 onto the Raspberry Pi. Be sure to remove the address jumper from the MCC 152. Restart the RPi and then use the DAQ HAT Manager or the Terminal command: sudo daqhats_read_eeproms Regards, Fausto
  19. Hello @lesthaeghet. Please confirm the acquisition input mode was changed to SingleEnded in QuickDAQ. What is the set sampling frequency? From the DT9844 User Manual, figure 7 shows a simple single-ended connection to a DT9844-32-STP box. Please test the following scenario with QuickDAQ and in single-ended mode. Disconnect all your wiring to the terminal blocks. Afterward, connect TB3, 18 (5V Analog) to TB1, 1 (Analog In 0). Also, connect TB3, 17 (Digital Ground) to TB1, 3 (Analog Ground) and TB3, 13 (Amplifier Low). Reply with a screen capture of the acquisition. The reading should be around 5V. If it is, retest your signal source. Regards, Fausto
  20. Hello @Dauple. Unfortunately, the DT9847-3-1 is no longer manufactured and an equivalent replacement is not available. Regards, Fausto
  21. Hello @Mike R. I am unable to find a NI PCI-6053E, but did find a NI PCI-6052E. Please confirm the NI device model. The NI PCI-6052E supports two analog output channels, which the MCC USB-1608G does not support. Please consider the MCC USB-1608GX-2AO. Since you and/or your colleagues are familiar with how the NI device is used and your application's requirements, compare those NI specs with the MCC USB-1608GX-2AO to determine if the MCC device is a suitable replacement. Note that your software application will need to be updated to support the MCC device. Regards, Fausto
  22. Hello @Ron Byron. Welcome to the Digilent Forum community. MCC related questions reside in the Measurement Computing (MCC) sub-forum. Use this link for the current version of MCC software for Windows, which includes InstaCal. The MCC software installation copies Universal Library examples to the C:\Users\Public\Documents\Measurement Computing\DAQ directory on the target system. Recommend using the UL Help as a reference for the supported functions of each device. An updated firmware version is available for the USB-1608GX via this link, along with instructions. Two versions are available depending on the hardware revision of the USB-1608GX. Select the firmware update compatible with your hardware revision. A label on the device identifies the revision. The letter just before the dash denotes the hardware revision. For example, part number 151440E-xxL indicates revision E. Regards, Fausto
  23. Hello @JMG. I am unable to assist on-site, but can attempt to troubleshoot your issue via this forum given additional information from you. Which MCC device, software application (and version), and operating system (and version) are you using? If you are unsure of the device model, please connect the device to a Windows system, launch MCC's InstaCal utility, and capture a screen image of the device information. Alternatively, snap an image of the physical device (top and bottom) and include it with your reply. Can you describe in detail (and capture screen images) of the issue(s) you are experiencing? Regards, Fausto
  24. Hello @Rob Clements. To be clear, you can drop a 'DT AIN00' module onto a worksheet, but the 'Select Board' field is empty. Is that correct? Open your Windows' Device Manager and search for a 'DT-Open Layers Data Acquisition Devices' branch. Does your DT9816 appear in that branch? If it does double click on the device to view the status on the General tab. What is the status? If you are using a virtual machine with Windows OS or Parallels Desktop on a MAC, then support is not available. If that's not the case, is the DT9816 plugged directly into a USB port on the system or is a USB hub (non-powered or externally powered) used? Externally powered USB hubs are preferred over non-powered USB hubs. Please reply with a screen capture of the expanded 'C:\Program Files (x86)\Data Translation' directory including the sub-folders. Next, with DASYLab not running, are you able to use the DT9816 device in Data Translation's QuickDAQ application? Lastly, please test the DT9816 on a different Windows system that does not have the same configuration or security settings as those machines experiencing an issue with the DT9816 device. Regards, Fausto
×
×