Dear all,

There is a customer who uses the USB-1608G with the underwater 4 units of Hydrophone TC4013.

They set the voltage to +1 voltage with DIFF mode on DAQami.

They found the voltage of the four channels is "-1" when they only connect the four TC4013 if they only select CH0~CH3 on DAQami.

The voltage becomes normal (+1 voltage) if they select all 8 channels (CH0~CH7).

Is the voltage normal?

 

And,

The user asks if they need to connect the load on the channels or not on the unused channels to let all the voltage of the channels with +1 voltage.

If it is needed, could you suggest how to set it up?

Thank you

Link to comment
Share on other sites

2 answers to this question

Recommended Posts

  • 0

Hello @Roger Lo.

Please confirm DAQami is set to Differential mode.

image.png

 

Select only those channels with a sensor connection.

image.png

 

Please verify the wiring connections to the USB-1608G analog input terminals are for differential wiring.  Add 100 kΩ resistors between CH#L and AGND terminals.  

image.png

 

Is the customer using a preamplifier with the TC4013 sensor?