jaejoo14 Posted November 7, 2023 Share Posted November 7, 2023 (edited) Scope Channel #1 is not working in Analog Discovery3. The same configuration as Channel #2, it has a very weak signal. Calibrated the AWG #1 and running the calibration of scope brings the exception because of weak signal. Any way to resolve it? Edited November 7, 2023 by jaejoo14 Link to comment Share on other sites More sharing options...
0 attila Posted November 7, 2023 Share Posted November 7, 2023 Hi @jaejoo14 1. Make sure the attenuation is set to 1 X 2. The input range relay may be stuck or remain in bad position due to certain software settings. Try toggling Channel 1 Range between 500 mV/div and 1V/div Does it solve the problem ? Can you hear it clicking like it is for Channel 2 ? 3. If you have calibrated the device (WaveForms/Settings/Device Manager/Calibrate), make sure the Scope 1 Gains are low values, or select Reset/Load Factory. Link to comment Share on other sites More sharing options...
0 jaejoo14 Posted November 7, 2023 Author Share Posted November 7, 2023 after reset and factory load. did AWG1 calibrated. While calibration of Scope Channel #1., Channel Reference Average Minimum Maximum CH1 0 V 0.25203 V 0.18282 V 0.32756 V Range: 62.404 V Offset: 0 V CH2 0 V 0.18155 V 0.11427 V 0.25138 V Range: 62.404 V Offset: 0 V Channel Reference Average Minimum Maximum CH1 0 V -0.60933 V -0.68159 V -0.53305 V Range: 62.404 V Offset: 19.997 V CH2 0 V -0.69062 V -0.76158 V -0.60542 V Range: 62.404 V Offset: 19.997 V Channel Reference Average Minimum Maximum CH1 0 V 1.1324 V 1.0549 V 1.211 V Range: 62.404 V Offset: -19.99663 V CH2 0 V 1.0743 V 1.0053 V 1.1539 V Range: 62.404 V Offset: -19.99663 V Channel Reference Average Minimum Maximum CH1 0 V 0.02201 V 3.3237 mV 0.04221 V Range: 5.4455 V Offset: 0 V CH2 0 V 0.01592 V 6.9797 mV 0.02526 V Range: 5.4455 V Offset: 0 V Channel Reference Average Minimum Maximum CH1 0 V -0.06411 V -0.0845 V -0.04329 V Range: 5.4455 V Offset: 1.9998 V CH2 0 V -0.07145 V -0.08085 V -0.06124 V Range: 5.4455 V Offset: 1.9998 V Channel Reference Average Minimum Maximum CH1 0 V 0.11067 V 0.08949 V 0.13669 V Range: 5.4455 V Offset: -1.99977 V CH2 0 V 0.10533 V 0.09481 V 0.11641 V Range: 5.4455 V Offset: -1.99977 V Channel Reference Average Minimum Maximum CH1 5 V 0.75628 V 0.66655 V 0.83414 V Range: 62.404 V Offset: 0 V CH2 5 V 5.9706 V 5.9037 V 6.037 V Range: 62.404 V Offset: 0 V Channel Reference Average Minimum Maximum CH1 -5 V -0.30918 V -0.3885 V -0.22472 V Range: 62.404 V Offset: 0 V CH2 -5 V -5.60836 V -5.6828 V -5.53425 V Range: 62.404 V Offset: 0 V Channel Reference Average Minimum Maximum CH1 2 V 0.23338 V 0.22534 V 0.2413 V Range: 5.4455 V Offset: 0 V Link to comment Share on other sites More sharing options...
0 attila Posted November 7, 2023 Share Posted November 7, 2023 Hi @jaejoo14 2 hours ago, attila said: 2. The input range relay may be stuck or remain in bad position due to certain software settings. Try toggling Channel 1 Range between 500 mV/div and 1V/div Does it solve the problem ? Can you hear it clicking like it is for Channel 2 ? For warranty and replacement contact the support.digilent @ ni.com specifying the Date of Purchase, Seller and Purchase Order/ Web order Number. Link to comment Share on other sites More sharing options...
0 jaejoo14 Posted November 8, 2023 Author Share Posted November 8, 2023 Yes, I can hear the clicking sounds for Channel #1 and #2. Still issue. Do you think I need to contact the support for a warranty? Link to comment Share on other sites More sharing options...
0 attila Posted November 8, 2023 Share Posted November 8, 2023 Hi @jaejoo14 Contact the support for replacement. The relay on oscilloscope channel 1 is stuck on high range, so the reading is probably correct only with 1V/div and higher. Toggling the relay a few times, like changing between 500mV/div and 1V/div, should have removed temporary stuck, contact issue, oxidation. You could try the following to make sure all application settings are reset, but don't expect to help. /Applications/WaveForms.app/Contents/MacOS/WaveForms -safe-mode (For nicer app interface you could use Settings/ Options/ Style : Fusion) Link to comment Share on other sites More sharing options...
Question
jaejoo14
Scope Channel #1 is not working in Analog Discovery3. The same configuration as Channel #2, it has a very weak signal. Calibrated the AWG #1 and running the calibration of scope brings the exception because of weak signal.
Any way to resolve it?
Link to comment
Share on other sites
5 answers to this question