ADP5250 Noisy AWG


JimR2

Question

Hi,

I just received my ADP5250, and was doing simple initial measurements on all the inputs/outputs, and am seeing significant noise (~500mV) on the AWG output. See below, is a 2V sine from the ADP5250 AWG, measured on a AD2. Regardless of cable or measurement device, the noise is consistent.

Also included is the wavegen screen of the ADP5250.

Any suggestions? I've got 2 AD2's, and a ADP3450, and have had no issues with the wavegen on those, this seems to maybe be a bad AWG? There's not much on the forum I could find on the 5250, so trying figure out whether I should send this back?

Also should add that this noise is seen, even with the AWG not running. In fact, I can turn off the ADP5250 while monitoring with the AD2, then turn back on the ADP5250, and watch the AWG output slowly float from ~0 to 500mV, at which point the noise disappears. Then when I turn on the AWG, the waveform looks reasonable (see third plot) but has a 500mV offset? Then, over time the noise re-appears, and the waveform moves back to that seen in pic#1

Regards, JimR2

  image.thumb.png.29bbdebd0a248433c760c4a266e8eaaa.png

 

image.png.5904c28894cead6ada21e7a882f28e6f.png4

 

image.png.d0eb1f79036c192e485cd9fdf18945a4.png

 

image.png.2b2461dc127b2bba33d51a55930a2203.png

Edited by JimR2
Link to comment
Share on other sites

3 answers to this question

Recommended Posts

  • 0

Hi @JimR2

First test the loopback using the ADP5250 by connecting the AWG to a Scope input. Is the signal clear ?

When connecting the ADP5250 AWG to the AD2 scope, are you using a BNC cable or a probe ? If not, is there a ground connection between the devices and is the negative scope input grounded ?

Link to comment
Share on other sites

  • 0

Hi Attila,

 

I did the loopback you mention to the ADP5250, actually was the first thing I did. I used three different BNC shielded cables, just to be sure. The signal was noisy in all cases.

The AD2 connection also used a BNC cable. I should have mentioned I used the BNC adapter with the AD2, so there was a ground

I did not try a scope probe, it was always a BNC cable direct connection.

 

Couple thoughts,

1) I got thinking maybe the AWG output is high impedance, and needs a common mode established by the load? 

2) There is a ground terminal on the back of the unit, should I be grounding that to the same ground as the AC supply to avoid ground loops? Didn't seem like this would be necessary for a local loopback check which I did.

 

Thanks for the reply, and look forward to working through this. Hope I'm doing something wrong, seems like a nice unit, although I'm also wondering about the packaging. The unit is not in a ESD bag when it comes in the box, and it has two Styrofoam supports which slip on over each end. I noticed one of the ends completely covers the AWG and Trigger ports comes in direct contact, seems like an ESD vunerability?

 

Regards, Jim Riches

Link to comment
Share on other sites

  • 0

Hi @JimR2

If the same problem persists with a loopback cable we can rule out external factors.
It could be an internal device fault on supply rails, reference... a cold solder or contact issue.

For warranty and replacement contact the support.digilent @ ni.com specifying the Date of Purchase, Seller and Purchase Order/ Web order Number.