Sangkyu Choi Posted April 14, 2023 Share Posted April 14, 2023 Hi! I'm going to measure impedance with AD2. I used W1-C1-DUT-C2-R-GND. I don't know is when measuring the circuit like below, I don't know if I should connect the reference resistor to GND or scope2(V-). Which of the 2 methods is correct? Link to comment Share on other sites More sharing options...
0 Anthocyanina Posted April 14, 2023 Share Posted April 14, 2023 there is a connection diagram this shows how to connect it, 1+ and 2+ are the scope channels, w1 is the generator output, and you see 1-, 2- (the negative of the scope channels) and ground are tied together Sangkyu Choi 1 Link to comment Share on other sites More sharing options...
0 attila Posted April 14, 2023 Share Posted April 14, 2023 Hi @Sangkyu Choi The negative oscilloscope inputs must always be connected to gnd, unless otherwise indicated, they are not used for differential measurements. Sangkyu Choi 1 Link to comment Share on other sites More sharing options...
0 Sangkyu Choi Posted April 17, 2023 Author Share Posted April 17, 2023 (edited) On 4/14/2023 at 3:41 PM, Anthocyanina said: there is a connection diagram this shows how to connect it, 1+ and 2+ are the scope channels, w1 is the generator output, and you see 1-, 2- (the negative of the scope channels) and ground are tied together On 4/14/2023 at 6:46 PM, attila said: Hi @Sangkyu Choi The negative oscilloscope inputs must always be connected to gnd, unless otherwise indicated, they are not used for differential measurements. Thank you! I'm sorry, I think I explained it wrong. Connectd is V+ is ch1+ v- is ch2+ ch1-, ch2- gnd The resistance I want to measure is 1 kohm, so 1 kohm should come out when measuring impedance. However, the impedance measurement in two ways did not produce 1 kohm. Can you tell me what the problem is? Edited April 17, 2023 by Sangkyu Choi Link to comment Share on other sites More sharing options...
0 attila Posted April 17, 2023 Share Posted April 17, 2023 Hi @Sangkyu Choi This is what I measure on a 1k 5% (DUT) with IA adapter which uses the same DUT-ReferenceResistor setup: Sangkyu Choi 1 Link to comment Share on other sites More sharing options...
Question
Sangkyu Choi
Hi!
I'm going to measure impedance with AD2.
I used W1-C1-DUT-C2-R-GND.
I don't know is when measuring the circuit like below, I don't know if I should connect the reference resistor to GND or scope2(V-).
Which of the 2 methods is correct?
Link to comment
Share on other sites
4 answers to this question