Recording faulty


Himayala1

Question

Hello,

I am using an analog discovery. When I use the record function it gives me measurment values with an offset error and a gain error. +100mV

The normal scope values basically differ from the record values.

Why is it doing this? How can I change my record values, meaning: rectify offset and gain error, so I get good values?

 

Edited by Himayala1
Link to comment
Share on other sites

8 answers to this question

Recommended Posts

  • 0

Hi @Himayala1

Are you comparing captures with the same rate ? At different rates higher frequency components may be sample/averaged differently.

Here I compared the following which overlap, except the sub-sample trigger positioning which is not applied for the imported data.
C1 - Rec. 200@50Hz 64bit + Import Bin
R1 - Record 200@50Hz
R2 - Single 4sec@8kHz

image.png

 

Link to comment
Share on other sites

  • 0

Hi @Himayala1

The 16bit format samples are raw samples.
To convert these to voltage see the note in the record dialog, like: V = 59.1970011866 * 1372  / 65536 - 0.002321186682
These range and offset values depend on the device/channel calibration.

The 16bit format may be useful for long records with binary, wav, tdms saving to reduce the file size.
The 32-64bit float values represent voltage, like 1.1 is 1.1V, these do not need conversion like the 16bit format.

image.png

Edited by attila
Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
×
×
  • Create New...