zja62 Posted July 14, 2022 Share Posted July 14, 2022 Hello, I would like to walk an example for the expected accuracy worst case on the AD2. Listed accuracy (scale ≥1V/div, VinCM = 0V): +/-100mV +/-0.5% So 1V/div, would be 1V +/- .105V or 10.5%? Likewise, Listed accuracy (scale ≤0.5V/div, VinCM = 0V): +/- 10 mV+/-0.5% So .1V/div, would be .1V +/-.0105V or 10.5% Please correct my understanding if I don't have this right. Link to comment Share on other sites More sharing options...
0 JColvin Posted July 14, 2022 Share Posted July 14, 2022 Hi @zja62, It is +/- 10 mV offset error and +/- 0.5% of the actual value. So for 1V, it would be between 0.985 V to 1.015 V. There are some more details about this same topic on this Forum thread here: Let me know if you have any questions. Thanks, JColvin Link to comment Share on other sites More sharing options...
Question
zja62
Hello,
I would like to walk an example for the expected accuracy worst case on the AD2.
Listed accuracy (scale ≥1V/div, VinCM = 0V): +/-100mV +/-0.5%
So 1V/div, would be 1V +/- .105V or 10.5%?
Likewise,
Listed accuracy (scale ≤0.5V/div, VinCM = 0V): +/- 10 mV+/-0.5%
So .1V/div, would be .1V +/-.0105V or 10.5%
Please correct my understanding if I don't have this right.