Letter of Volatility for Analog Discovery 2


nkornbau

Question

Hello,

We have purchased some Analog Discovery 2 devices for some testing we are doing, and would like to be able to clear or reset any/all memory on board to the original state. Can Digilent provide a Letter of Volatility for the Analog Discovery 2?

Thanks!

Nathan

Link to comment
Share on other sites

2 answers to this question

Recommended Posts

Hi @nkornbau

There is no dedicated volatile memory in Analog Discovery 2, only the registers/BRAM in C36 (FPGA), registers in IC32 (Clock Generator), IC10 (DAC), IC12 (ADC), IC14 (DAC), IC43 (DAC), IC21,26 (Power Monitor), IC45 (Temperature Sensor).

The only non-volatile memories in the device are: IC37 (storing USB IDs, SN, device name) and IC39 (factory and user calibration data) EEPROMs.

The non-volatile user data can be changed and erased/overwritten here:
- user calibration data can be changed under WaveForms application Settings/ Device Manager/ Calibrate/, erased with Reset/ Load factory or Reset to zero option
- device name edited/overwritten under WaveForms application Settings/ Device Manager/ Rename 

I've asked my colleague to send an official letter to your email address.

Link to comment
Share on other sites