  1. Hi @brandonhill7

    The average and oversampling needs stable and repetitive trigger. The acquisition will stop if there are no more trigger events and the trigger is set to normal.

    To prevent the FFT phase alignment to be used, reduce the number of samples to have less than 8 periods, like for 158kHz signal ~4ki samples, or set the hidden option Scope.Trigger.Phase.text = "Xing" for alignment based on trigger crossing interpolation.

    image.png

  2. Hi @Mike4354

    Here the Logic Analyzer detector is set to trigger on DIO 24 falling edge and Pattern Generator outputs a 100ns low pulse on DIO 25 with 200ns (20 cycles @ 100MHz) + Wait time delay.

    image.png

     

    You could adjust the system frequency to change the delay, 250ns (20 @ 80MHz) +

    image.png

     

    With ROM logic the delay is lower, 80ns at 100MHz (DIO 26 is used as state bit)

    image.png

    image.png

     

    Reducing the ROM frequency will result in delay uncertainty, but delay could be achieved with further state bits.

     

  3. Hi @SpinWizard

    1. Adjust the format, channels and frequency according your hardware specs or requirements. The autodetection of this is not working due to the software abstraction layers.
    2. Higher voltage rating would require different protection, certification, costs.... 10x probe increases the voltage range but above 50V with 1x can damage the device.
    3. I'm not sure I understand what you are asking for. Do you need digital multimeter with USB and software support ?

  4. Hi @fr.bernardo

    You can change the CS in SPI/Custom, but make sure to have pullup on these keep the unused ones idle, or you can control from Script tool the SPI and StaticIO.
    You could also have common CS/CLK and separate MISO, supported only by SPI/Spy. In SPI/Custom separate calls are needed like with different CS, eventually calling ReadQuad for 4 sensors and unpacking the bits.

    image.png

  5. In Network Analyzer you can set Source: External and control the sweep from Wavegen or external equipment.
    The NA will look for peak in the spectrum and consider this the signal frequency. Under Options menu you can adjust the External Threshold, the minimum peak magnitude.
    Depending or number of Steps, set the sweep time to seconds or minutes, for wide logarithmic scales you can adjust the sweep frequencies manually. When all the step points are captured it will draw a line, until this only dots.

  6. Hi @gcb

    In Pattern Generator the DIOs can be configured to generate different frequencies, pulses, phases... optionally each having 1k-32k custom pattern buffer. In a bus the selected DIOs are configured at sequential phase, period... settings to generate various types of counters.
    The Play mode Logic Analyzer is different, currently only available with Digital Discovery. Main usage is to regenerate the capture on 1 or more DIOs but it can also generate data from imported files, DIOs have at same frequency and phase but different bit positions of samples.

    The application lets you open multiple instruments (Scope, Wavegen, Logic, Patterns, Logger, Voltmeter, Spectrum, Network, Impedance Analyzer, Tracer...) relying on the same device resource (Scope, Wavegen channels, Logic, Patterns, Supplies, StaticIO). Only one of these instances can control at a time a device resource, the last used instance (Run or Stop press), the other display Busy state.
    Like having multiple identical interfaces open only the last started one will capture, the other will be busy.
    The Voltmeter, Logger and Spectrum also uses the Scope device resource.
    The Wavegen channels are separate device resources, can run independently, except when used in Synchronized mode.
    The Network and Impedance Analyzer use Scope and Wavegen 1 channel, by default, also Supplies and StaticIO when used with IA adapter.
    The Tracer uses Scope, Wavegen 1 and 2, also Power Supplies and StaticIO when used with Transistor tester adapter.
    The Protocol tool uses Patterns device resource to send and Logic to receive data.
    The Logic uses the Patterns resource in Play mode.

    Regarding your junk.txt which is actually a csv, change the "List separator" to comma.
    The default delimiter for TXT files is Tab and for CSV is comma. This is regional system option dependent but can be set to US under Settings/Options/Application.

    image.png

     

    If you want to generate burst like in your file, this can be done with all T&M devices and Pattern Generator.

    image.png

     

    The demo mode is intended to explore the application and device features, but it does to fully emulate all the features.

    image.png

    image.png

    DIO24 wired to DIN0

    image.png

     

     

  7. Hi @StefKay

    The channel indices up to 2x are Filter channels and the following ones are Wavegen digital readbacks.
    Like AD3 has two real channels + 2 filter channels + 2 AWG readbacks. Indices 0 and 1 are for the scope channels, 2 and 3 are for the filter channels, 4 and 5 are for wavegen readbacks.
    The EcplyseZ7 and ADP3X50 has additional readback for FM/PM and AM/SUM nodes.

    Enable these with FDwfAnalogInChannelEnableSet. Note, the device capture memory is shared between the enabled (1,2,4,8) channels so the depth decreases as more are enabled.
    Like AD3 with the first default configuration has a total of 32k samples buffer. Having one channel enabled you get 32k sample, with 2 enabled 16kS/channel, with 3 or 4 8kS/ch and with 5 or 6 4kS/ch.
    The sample applies for EcplyseZ7 and ADP3X50 having 256M and 128M sample memory.

  8. Hi @StefKay

    The FDwfEnumUserName returns the editable name or use FDwfDeviceOpenEx("name:mydevicename\0", &hdwf)

    The ElypseZ7 is a development board. The WaveForms support for it originally was only used for internal Zmod development, validation, testing and calibration. The board's SN and device name are stored in the other USB controller's EEPROM and the reserved fields for these in the QSPI are empty.
    To change these, in WaveForms hold down the Ctrl key while opening Settings/'Device Manager'. Now under More dropdown a 'Device IDs' option will show up where you can change the names and SN by entering new value, and press the Set button after the field, but don't change the Prod|Rev|FW.

    image.png

  9. Hi @scorbetta

    The rgValue and rgFlag expects 32bit unsigned values.
    You should store the 'read' as an ID, for instance flag value 1 to represent 'read', 2 to represent 'write'...

    function Value2Text(flag, value){
      switch(flag){
        case 0: return "nop";
        case 1: return "RD b"+value.toString(2);
        case 2: return "WR h"+value.toString(16);
        default: return "DAT "+value;
      }
    }

×
×