Abhijan reacted to attila in Pass Oscilloscope Value to Waveform Generator
Hi @Abhijan
In-device Oscilloscope input to AWG output routing (+amplification, filtering) is only supported by newer devices AD3, ADP3X50...
With AD3 this adds a delay of about 20 cycles (200ns at 100MHz) or more with filtering.
-
Abhijan reacted to JColvin in Waveforms Voltage Level Does Not Match Measurement
Hi @Abhijan,
The first thing I would be checking with the 1:10 ratio would be the attenuation setting on the BNC probe. It's also worth ensuring that your two devices share the same ground connection.
Let me know if you my guess is not accurate to your situation.
Thanks,
JColvin