Hello,
I have a type s thermocouple that is connected to the MCC USB-2001-TC unit, the thermocouple goes into an oven that can go up to 2700F. I Use LabVIEW to collect the data from the DAQ. I've tried to collect the temperature readings but whenever the temperature reaches ~1700F, the readings drop down to ~1400F and stay there until the temperature drops back down below 1700F. However, I tried to collect the raw voltage from the TC and those values seem to follow the temperature better, so I'm trying to convert the raw data to temperature but can't seem to find a way to do that. I've tried to follow the NIST Type S coefficient but the raw data is ranging between 632500.3 at 2700F and 524338 at 80F, which is way higher than the NIST mV ranges. Is there a way I can convert the raw data to temperatures?
Thank you,