AndyHook1970 Posted July 20, 2017 Share Posted July 20, 2017 Just two quick questions from a newbie. 1) Will I be able to measure on the Scope a 100VDC signal that is derived from mains? I plan to use a 10:1 probe. 2) Can I cause any damage by exceeding the 25V input limit on the analogue inputs? Many Thanks Andy Link to comment Share on other sites More sharing options...
attila Posted July 21, 2017 Share Posted July 21, 2017 Hi @AndyHook1970 2. The recommended maximum voltage for scope inputs of Analog Discovery (1&2) is ±25V (50V differential) and the absolute maximum is ±50V. Exceeding the abs max could damage the device. See:https://reference.digilentinc.com/reference/instrumentation/analog-discovery-2/reference-manual#scope_input_divider_and_gain_selection 1. With 10x probe you could measure up to ±250V, but be cautions when working with high voltages. To see the real value, under channel options configure the applied attenuation: Link to comment Share on other sites More sharing options...
Question
AndyHook1970
Just two quick questions from a newbie.
1) Will I be able to measure on the Scope a 100VDC signal that is derived from mains? I plan to use a 10:1 probe.
2) Can I cause any damage by exceeding the 25V input limit on the analogue inputs?
Many Thanks
Andy