t991731 Posted December 6, 2016 Share Posted December 6, 2016 Hi! I had problems with analog discovery 2. I suspected that I erased the eeprom of the FTDI and after restoring it with the utility (thank you, attila) the device complained of being uncalibrated. When I tried to calibrate it the 0v level the measured value in the first step of calibration of W1 was 126mV and the value for 1V was 130mV. When I started to investigate how the other functions perform then I discovered that the V+ output gives 5.38V, -5V output gives ~0V. The waveform generator outputs the selected signals but the amplitudes are wrong, instead 1vp-p I get signal between 100mV and 200mV. So it seems like the internal negative voltage has failed. I inspected the board and I didn't see any visual faults. What would be the best course to start troubleshooting and looking for the error? With best regards, Marek Link to comment Share on other sites More sharing options...
attila Posted December 7, 2016 Share Posted December 7, 2016 Hello, In the calibration window under Reset menu try Load factory or Reset to zero. Link to comment Share on other sites More sharing options...
t991731 Posted December 8, 2016 Author Share Posted December 8, 2016 Hi! Thank you for replying. I tried both but this did not solve things. Still when I generate 1Vpp sine in the generator I see only 50mVpp sine with the scope. The V+ supply gives 5.38V and the scope measures it as 6V The V- supply gives ~50mV (both supplies are turned on) I also thought that pattern generator and logic analyzer wont work but they do but the funniest thing is that the pins are swapped. I configured DIO15 as output and sent a clock pattern to it and did not see any change in that pin. The pin to which that signal was routed was DIO10. Device is also recognized as Discovery. Not as Discovery 2. Thats the problem, i think. Now how to fix it? Regards, Marek Link to comment Share on other sites More sharing options...
t991731 Posted December 8, 2016 Author Share Posted December 8, 2016 I managed to fix it. I rewrote name "Discovery" in FT_Prog to "Discovery2" then rerun the digilent recovery tool and it worked. Strange that the recovery tool alone could not do the trick, I had run it at least 3 times before.... Link to comment Share on other sites More sharing options...
Question
t991731
Hi!
I had problems with analog discovery 2. I suspected that I erased the eeprom of the FTDI
and after restoring it with the utility (thank you, attila) the device complained of being uncalibrated.
When I tried to calibrate it the 0v level the measured value in the first step of calibration of W1 was 126mV
and the value for 1V was 130mV.
When I started to investigate how the other functions perform then I discovered that the V+ output gives 5.38V, -5V output gives ~0V.
The waveform generator outputs the selected signals but the amplitudes are wrong, instead 1vp-p I get signal between 100mV and 200mV.
So it seems like the internal negative voltage has failed. I inspected the board and I didn't see any visual faults.
What would be the best course to start troubleshooting and looking for the error?
With best regards,
Marek