Probe compensation for attentuation over frequency


Kirkhaan

Question

Hi forum members,

I am using an AD2 with a passive 1:10 voltage probe that has the following transfer function:

Screenshot.png

So as can been seen, its 1:10 attenuation drops over frequency.

Is there a way to compensate for this in using custom probe settings?
In WaveForms I only found Linear, Table and Custom tabs, which seem to be only voltage related.
So no dynamic attenuation possible over frequency.

 

Edited by Kirkhaan
Link to comment
Share on other sites

2 answers to this question

Recommended Posts

  • 0

Hi @Kirkhaan

First try to tune the probe.

image.png

 

The Mode refers to the stimulus signal amplitude.
For relative magnitude, save a trace as Reference (or import from file or generate it with Script...) and use Math like this:

image.png

Link to comment