I needed some clarification on the thermocouple measurement accuracy table footnote on page 14 of the USB-TC User's Guide
(https://www.mccdaq.com/PDFs/Manuals/usb-tc.pdf). The note says,
"Thermocouple specifications include linearization, cold-junction compensation and system noise. These specs are for one year, or 3000 operating hours, whichever comes first and for operation of the device between 15 °C and 35 °C. For measurements outside this range, add ±0.5 degree to the maximum error shown. There are CJC sensors on each side of the module. The accuracy listed above assumes the screw terminals are at the same temperature as the CJC sensor. Errors shown do not include inherent thermocouple error. Please contact your thermocouple supplier for details on the actual thermocouple error."
Does this mean that the measurements made with the USB-TC need ±0.5 degrees added to the maximum error after:
a year/3000 operating hours or
if the device is outside the 15°C and 35 °C range or
both 1. and 2.?
if the measurements need the added error range only for 2., what would the maximum error be after 1./3.?
Question
ChrisB
I needed some clarification on the thermocouple measurement accuracy table footnote on page 14 of the USB-TC User's Guide
(https://www.mccdaq.com/PDFs/Manuals/usb-tc.pdf). The note says,
"Thermocouple specifications include linearization, cold-junction compensation and system noise. These specs are for one year, or 3000 operating hours, whichever comes first and for operation of the device between 15 °C and 35 °C. For measurements outside this range, add ±0.5 degree to the maximum error shown. There are CJC sensors on each side of the module. The accuracy listed above assumes the screw terminals are at the same temperature as the CJC sensor. Errors shown do not include inherent thermocouple error. Please contact your thermocouple supplier for details on the actual thermocouple error."
Link to comment
Share on other sites
1 answer to this question
Recommended Posts
Create an account or sign in to comment
You need to be a member in order to leave a comment
Create an account
Sign up for a new account in our community. It's easy!
Register a new accountSign in
Already have an account? Sign in here.
Sign In Now