Don't see expected bandwidth in AD2


miv2k

Question

Hi,

I've been playing with Analog Discovery 2 (totally new to it, so please forgive me if I ask stupid questions) and noticed that when a 5MHz square wave from WG1 was scoped by a 1x BNC probe on CH1, it looked like a triangular wave. That was somewhat unexpected to me, since the specs call for 30+MHz bandwidth @ -3dB.

I then dug out an AD2 Reference Manual, and on page 30 of it saw how they demoed AWG spectral characteristics by running Network Analyzer in WaveForms by feeding WG1 directly to CH1, and their plot was also suggesting -0.5dB @ 5MHz and -3dB @ 20MHz.

I did the same, and got way different results: -0.5dB @ 2MHz and -3dB @ 5.5MHz for 1x probe and even more bizarre plot for 10x.

Why are my results so much worse than the specs? What may I be doing wrong here? All screenshots are attached.

Thanks!

 

This is how how a 5MHz square wave looks when scoped by enclosed BNC 1x probe:

5mhz_bnc_1x.png

This is characteristic measured by 1x BNC probe:

net_analyzer_bnc_1x.png

This is characteristic measured by 10x BNC probe:

net_analyzer_bnc_10x.png

This is what AD2 reference manual calls for when BNC probes are used ("Figure 21 shows the typical spectral characteristic of the AWG. In the first experiment (solid line), a coax cable and a Digilent Discovery BNC adapter were used to connect the AWG signal to the Scope inputs")

net_analyzer_expected.png

Edited by miv2k
Link to comment
Share on other sites

11 answers to this question

Recommended Posts

  • 0

Oh, you mean you skipped probes altogether and just used a BNC to BNC cable...

Here's my capture with 1x probe going to WG1 0R, and I think it matches your green plot.

Should I blame then the oscilloscope probes that came with AD2 for the reduced frequency range?

I then fail to understand marketing material... "When used with BNC probes the adapter increases the oscilloscope bandwidth from 9MHz to 30MHz". Even if we say that the bandwidth is given by -3dB cut-off (which for measuring equipment is kinda extreme), you and I are seeing only 8MHz with the BNC probes, far cry from 30MHz.

net_1x_0R.png

Edited by miv2k
Link to comment
Share on other sites

  • 0
  • 0

Thanks for taking your time to answer my questions. I don't have another waveform gen capable of 5-30MHz, but I'll see if I can come up with 5-10MHz square wave and check if it looks better on AD2 scope than the one in the first screenshot. If it does, then yeah, it will confirm that I'm hitting AD2's WG limits here, not the scope.

Link to comment
Share on other sites

  • 0

IDK. What could be the reason?

Initially I thought that ringing could be because I just hacked it on a breadboard with long leads on the feedback resistor, but now I fed the generated signal to another trigger in the same chip and scoped right from the second trigger's output pin. I still see ringing. Higher frequency too now. Probably because I'm not loading feedback with the probe's capacitance anymore.

 

6mhz_74hc14n_bnc_10x_buffered.png

Edited by miv2k
Link to comment
Share on other sites

  • 0

Well, the ringing as a concept is no secret, but what's causing it here? Inductance in breadboard? Too fast a transition?

For the heck of it I removed crocodile ground wire from the probe and added a thing like below, but it didn't change the situation much.

 image.png.b7ad24d17e74763e713e68309aa7e51c.png

Link to comment
Share on other sites