I've been playing with Analog Discovery 2 (totally new to it, so please forgive me if I ask stupid questions) and noticed that when a 5MHz square wave from WG1 was scoped by a 1x BNC probe on CH1, it looked like a triangular wave. That was somewhat unexpected to me, since the specs call for 30+MHz bandwidth @ -3dB.
I then dug out an AD2 Reference Manual, and on page 30 of it saw how they demoed AWG spectral characteristics by running Network Analyzer in WaveForms by feeding WG1 directly to CH1, and their plot was also suggesting -0.5dB @ 5MHz and -3dB @ 20MHz.
I did the same, and got way different results: -0.5dB @ 2MHz and -3dB @ 5.5MHz for 1x probe and even more bizarre plot for 10x.
Why are my results so much worse than the specs? What may I be doing wrong here? All screenshots are attached.
Thanks!
This is how how a 5MHz square wave looks when scoped by enclosed BNC 1x probe:
This is characteristic measured by 1x BNC probe:
This is characteristic measured by 10x BNC probe:
This is what AD2 reference manual calls for when BNC probes are used ("Figure 21 shows the typical spectral characteristic of the AWG. In the first experiment (solid line), a coax cable and a Digilent Discovery BNC adapter were used to connect the AWG signal to the Scope inputs")
Question
miv2k
Hi,
I've been playing with Analog Discovery 2 (totally new to it, so please forgive me if I ask stupid questions) and noticed that when a 5MHz square wave from WG1 was scoped by a 1x BNC probe on CH1, it looked like a triangular wave. That was somewhat unexpected to me, since the specs call for 30+MHz bandwidth @ -3dB.
I then dug out an AD2 Reference Manual, and on page 30 of it saw how they demoed AWG spectral characteristics by running Network Analyzer in WaveForms by feeding WG1 directly to CH1, and their plot was also suggesting -0.5dB @ 5MHz and -3dB @ 20MHz.
I did the same, and got way different results: -0.5dB @ 2MHz and -3dB @ 5.5MHz for 1x probe and even more bizarre plot for 10x.
Why are my results so much worse than the specs? What may I be doing wrong here? All screenshots are attached.
Thanks!
This is how how a 5MHz square wave looks when scoped by enclosed BNC 1x probe:
This is characteristic measured by 1x BNC probe:
This is characteristic measured by 10x BNC probe:
This is what AD2 reference manual calls for when BNC probes are used ("Figure 21 shows the typical spectral characteristic of the AWG. In the first experiment (solid line), a coax cable and a Digilent Discovery BNC adapter were used to connect the AWG signal to the Scope inputs")
Link to comment
Share on other sites
11 answers to this question