Hello, my question is about the USB-1616HS. I mounted 2 thermocouples on it. When I take the same temperature at a same point multiple times (I am using TracerDAQ), the temperature detected is always different than the previous one and it is not the expected temperature (it was supposed to be 115 degrees in both graphs below).

I wanted to know if there is something I am not doing the right way as the temperature should be the same between every measurement. 

Thank you for your answer,

Louis R.

 

image.png.be9651065c9ac195d98a2b9d51d78131.pngimage.png.4a0d6148a7e7b3efe2487eeeda254804.pngimage.png.0ff1ccce26e8713667cc50582528bca9.png

Link to comment
Share on other sites

5 answers to this question

Recommended Posts

  • 0

Hello @LouisR.

What is the age of your USB-1616HS module?

Have you acquired correct temperature readings previously, with or without using TracerDAQ, and with those specific thermocouples?

Have you tested other thermocouples with the USB-1616HS module?

Are the ends of the thermocouples connected directly to a surface or in the air?

If you exit TracerDAQ and open InstaCal to take some measurements, are the readings correct?

Regards,

Fausto

 

Link to comment
Share on other sites

  • 0

I don't know the exact age of my USB-1616HS, the serial number is 394943 (someone told me it has been at least 3 years)

I never worked on this module or these thermocouples before so I have never acquired the correct temperature using it.

I currently only have these 2 thermocouples so I didn't try with another one. 

The ends are connected directly to a surface. 

And about Instacal, I used it to calibrate the device before utilization but the results still aren't accurate.

Thanks for your help,

Louis R.

Link to comment
Share on other sites

  • 0

Hello @LouisR.

Your serial number dates back to May 2019.

Your image of the USB-1616HS and thermocouple show a thin film tape at the thermocouple's tip.  Please confirm that both thermocouple tips are not in direct contact with a metal surface.  Also, please verify that the thermocouple leads are secured at the screw terminals.  Reference Figure 1 on page 9 of the USB-1616HS User's Guide for a thermocouple connection.  Your thermocouple connections should begin with analog input channel 0 and you should not leave an open channel between thermocouple connections.

https://www.mccdaq.com/pdfs/manuals/USB-1616HS.pdf

In InstaCal, verify that the Input Mode is set to 'Differential (8ch)'.  

image.png

 

Configure the thermocouple channels on the next tab.

 

image.png

 

What are the lengths of your thermocouples?

Additional 'tips for making accurate temperature measurements' are listed on pages 19-20 of the USB-1616HS User's Guide.

If you continue to have an issue with your temperature measurements, then we can take this offline to discuss a RMA for sending the device in to be tested/repaired. 

Regards,

Fausto

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
×
×
  • Create New...