Hi @sdrake It would be useful if you could post a screenshot... 1. In case the measured period is not multiple of the signal frequency the average will be high, depending on the captured sine periods and the additional fraction. 2. You can use the Zero offset to eliminate device and external circuit DC error. 3. You can recalibrate the device under Settings/ Device Manager. Here C1 captures 2 periods and C2 1.5 :   Perform capture having inputs at GND or circuit 0V (