Hello
I certainly miss something, but I cannot figure out how to trigger pattern with an AWG output.
According to the reference manual: Each instrument generates a trigger signal when a trigger condition is met.
It's rather easy to trigger the AWG and the pattern output with the scope, but I didn't figure out how to trigger patterm output with AWG.
I certainly didn't found what a "trigger condition" is for an AWG.
What I'm trying to do is to have two AWGs synchonized and on
Hi @GCL
The instruments output trigger high while running. Selecting Wavegen as trigger output it will constantly output high after starter, if no trigger condition or wait time is set.
Your approach with scope and patterns looks good, but in order output pulse with each trigger event select "scope detector" as trigger in patterns. In this case you don't have to have the scope running. You can create signal with constant, like 0 in idle and output high, with wait and run you can adjust