USB-2416-4AO DI/O DAQAMI SUPPORT


smann

Question

Hi I am using the DAQ for measuring a pressure transmitter. Even at no pressure the DAQAMI software is reading unstable values when in theory it should be at constant 0 at atmospheric pressure. Any ideas on how I can get a constant reading from the DAQ, I confirmed it is wired correctly . image.thumb.png.a8b802f7bef459486e38f3f26d3f34b1.png

Link to comment
Share on other sites

1 answer to this question

Recommended Posts

  • 0

Hello @smann.

In DAQami, enable analog input channel 0 and configure it to differential mode.  Set the data rate to 60 Hz.  Connect a jumper wire between terminals CH0H and CH0L.  Start an acquisition in DAQami to determine the baseline noise level for your USB-2416-4AO.

Connect your signals beginning with analog input channel 0 and only enable those channels intended for capturing data.

Regards,

Fausto

Link to comment