Hi John,
Thank you for coming back to me.
We have found this issue on many examples of the DT9805 old and new.
Running a long data acquisition. Stopping it, then starting another one is the most likely scenario to cause the problem, but it does equally just occur mid data acquisition at random times.
Temperature is not an issue, we see the issue at normal room temperature i.e. around 20 degrees c. There are no covers on the instrument which is well ventilated, Although it is a fire test instrument I am running dummy tests so no heating is involved. In any case all electronics are well separated from heated areas.
After the USB has dropped, the calling application (our software or QuickDaq) cannot be closed from task manager.
A soft reboot does not cure the problem and the application is still shown in task manager as running
Only a hard reboot (press and hold the PC power button) closes the application. This to me is consistent with a driver crash/hang where the driver is waiting for a hardware action to complete.
My tests this time around were on a brand new unit after a customer reported this issue with a different (brand new) unit in the field.
It was me that reported the issue in 2019 and it has never been resolved since then, only now we are again getting high level complaints from customers who have been experiencing this issue for a while.