  1. We have attempted to configure a DT9844 to read 2 separate SE analog input channels and display the output as a function of time using LabVIEW.
    The read rate is set to 300klHz.
    On AI0 we have an input with V~=2.6V (constant).
    On AI1 we have an input V nominally at 2.27V that can increase to ~2.8V.
    Measuring the V at the AI terminals with an accurate precision DMM, we have confirmed that the V on AI0 is steady at 2.603V and that AI1 ranges from 2.27 to 2.7V.
    The V measured by the DT9044 on AI1 seems to accurate with respect to the DMM to within 1mV.

    However, as the V changes on AI1, a V change is produce on our AI0 display.  We suspect there is an AI configuration error.  Our sample code is show below and attached.  Any suggestions of where the configuration error is would be greatly appreciated.

    image.png.eed5e3c390173081f13d9f5b3806c7a5.png

     

    image.png

    DT_nCH_nSAMP_Cont.vi

  2. based on the configuration, minretriggerperiod should be:

     

    (1 CGL * 100 CGLs per trigger)/500kHz + 2us = .000202s

    This implies a max retrigger of 1/.000202s or 4950Hz.

    the configuration calls for a retrigger at 300Hz which should be supported, correct?

     

    If there is only 1 CGL (AI0) and samples per channel is set to 100 and scan count is 100, should there be only 100 samples collected per trigger?

  3. We are using an external trigger and an external clock to acquire a signal on AI0 of a DT9844.

    The external clock is a square wave, 50% duty cycle, 500KHz

    The external trigger is a square wave, 50% duty cycle, 300Hz.

    The AI0 input is a a square wave, 50% duty cycle, 150Hz, 0 to 5V

    AI0 is configured to scan the channel list (AI0 only) 100 times, to collect 100 samples per scan and to buffer a total of 10000 samples to display.

    We anticipated the displayed buffer to have 10000 samples, looking like a square wave of 100 samples at 5V followed by 100 samples of 0V repeated 50 times.

     

    We were able to configure an NI6343 to collect in this manner and the displayed output was what was expected.  The block diagram and output display from the NI6343 are show below.  The display shows the signal for one triggered scan and the complete buffer.

    NI6343AI0ScanandBufferBlockDiagram.thumb.png.88c836569e0e3b9dc988ccc4f3e31b5c.pngNI6343ScanandBuffer.thumb.png.78f75a837dd754a95190290788a733bd.png

     

    Also shown below are the block diagram for the DT9844 and the output display.  While the display for one triggered scan is what was expected, the complete buffer was not.

    DT9844AI0ScanandBufferBlockDiagram.thumb.png.9178c60d90f6117e7c43e712e2f39b11.png

     

    DT9844ScanandBuffer.thumb.png.54f8e77ae7cb50ca9e518ce145954545.png

    The number of samples per scan, scan count and buffer size may not be configured properly and are likely the cause for the 10000 buffered samples to seem to be random.

    It there an improper configuration for these parameter in the block diagram? 

    NI6343 Scan and Buffer.png

  4. When controlling a DT9844 from LABView, is there a way to configure the counters to produce a finite series of pulses each time a gate is measured?

     

    For example, if CTR0 is set as a gate based on a rising edge, if it sees a rising edge, can CTR1 be configured to produce say 5 pulses, each pulse being 50% duty cycle square wave with a 10ms period then stopping until the next rising edge is seen on CTR0?

    If CTR0 sees a 1Hz square wave, each time CTR0 goes hi, CTR1 would generate a 50ms output of a 100Hz, 50% duty cycle square wave followed by lo for 950ms.

     

    To cross reference, some NI products support Implicit Timing and Pause Triggers on Counters. See:

    https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019M9CSAU&OpenDocument=&l=en-US

    and

    https://www.ni.com/docs/en-US/bundle/ni-daqmx/page/mxcncpts/gensnglefinconttrains.html

×
×